    Hello.

    I had FSH4 with FW version ... it seems 1.41.
    but when I installed version 2.5 or 3.10, when booting appeared the warning:
    The Self Alignment Data is not compatible with the current Firmware Version.

    maybe this is a stupid question, but someone knows how to fix it ?

    I have the oldest firmware V2.50. maybe someone has an older one ?

    Thanks.

    1124201201a_HDR.jpg
    /carv

  2. # ADS
    Circuit advertisement
    Join Date
    Always
    Location
    Advertising world
    Posts
    Many
     

  3. #2
    VIP Member Reputation: 1852 carver's Avatar
    Join Date
    2007-10-24
    Location
    Ukraine
    Posts
    563


    1 out of 1 members found this post helpful.

    Default Re: R&S FSH4 (The Self Alignment Data is not compatible with the current FW)

    resolved.
    Setup -> Self alignment.
    but need FSH-Z28 (FSH-Z29)

    the only thing is that InstrumentView
    for some reason doesn't see the device over Ethernet.
    although I buy all the options for this fsh4.

    Installed option1e.jpg
    Last edited by carver; 2020-12-06 at 08:07 PM
    /carv